Hi,
Sorry to bother you with this, but I could not find a satisfactory answer when searching old forum posts.
I have 2 of the Velleman K8067 kits assembled and working, but need to know the actual correlation 0-20mA to centigrade mapping.
The specification states: -20 to +70C for 0-20mA.
So, with industrial sensors one clearly assumes that 0-20mA maps to -20 to 70C.
Therefore using 250R resistor (giving a 0-5V) the following equation should therefore hold:
degC = -20 + 18 * V
However the equation in the manual suggests a constant of -23. So do I assume that the actual range for 0-20mA is -23C to 67C, -23 to 70C or something else completely?
I don’t really want to have to resort to deriving a calibration curve since this would defeat the whole object of getting a properly tested kit using a nice linear output LM335 and TLV2741 rail-rail op-amp.
Initial results indicate that an offset of -30.6 and slope of 21 are closer to measured values using a Fluke 1587 and a type K thermocouple. i.e. this gives a range of -30.6 to 74C for the 0-20mA range.
Any clarity on this much appreciated.
Regards
Eclectica