Problem with PPS10 – faulty data from serial port

Dear Velleman Support,
I bought a new PPS10 this week from a dealer here in Germany. Measuring itself works fine. But unfortunately I get no correct measurement on the PC.
Today I did intensive testing to determine if there is a problem with my configuration or the PPS10.

First of all a description of my testing conditions:

  • Problem occurs on 2 different computers
  • Problem occurs on Windows 7 as well as Windows XP
  • Both Desktop PCs I’ve tested with have onboard Com Ports (No USB Adapters used)
  • PPS10 powered with alkaline batteries. Unfortunately I don’t have a matching external power supply at the moment
  • Firmware of the PPS10 is “May 2006 Build 080506F01” (is there a newer one ?)
  • tried to reset the PPS10. Doesn’t change anything.

During all the tests described below a square wave was measured. Here is what the Scope showed on the screen (all the time):

When I start the PC program, I get a reading but the graph is redrawn only every few seconds. It takes between 1s and 14s in a very irregular manner. Data is shown more sporadically than systematically. The settings are recognized correct except of the triggering. The Triggering field on the status bar shows always “norm” instead of “run”.
That few data shown differs between a clean measurement without much jittering and a totally jittering measurement.
Hereby the “clean” measurement has strange values. First I thought the data packages would be send in another order (newest samples first, oldest samples last). But the volt values themselves do not match the Curve on the Display of the scope.

The Pictures below show what the PPS10 Software looks like:

To examine the data arriving from the scope to the PC more detailed I’ve downloaded a Demo Version of a Terminal program: http://www.eltima.com/products/serial-port-monitor/

The data arrives absolutely continuous without any pauses. Here’s an extraction from the measurement:

42 41 0a 01 07 08 06 00 73 72 73 72 73 72 73 72 BA......srsrsrsr 73 72 73 72 73 72 73 72 73 72 73 72 73 72 73 72 srsrsrsrsrsrsrsr 73 72 73 72 73 72 73 72 73 72 73 72 73 72 73 72 srsrsrsrsrsrsrsr 73 72 73 72 73 72 73 72 73 72 73 72 72 72 73 72 srsrsrsrsrsrrrsr 73 72 73 72 73 72 73 72 73 72 73 72 73 72 73 72 srsrsrsrsrsrsrsr 72 72 72 5f f9 fd f9 fd f9 fd fd fd f9 fd f9 fd rrr_ùýùýùýýýùýùý fd fd fd fd f9 fd fd fd f9 fd f9 fd f9 fd fd fd ýýýýùýýýùýùýùýýý fd fd f9 fd fd fd f9 fd fd fd fd fd f9 fd fd fd ýýùýýýùýýýýýùýýý fd fd f9 fd fd fd f9 fd f9 fd fd fd f9 fd f9 fd ýýùýýýùýùýýýùýùý fd fd f9 fd f9 fd f9 fd f9 fd f9 fd f9 fd f9 fd ýýùýùýùýùýùýùýùý f9 fd fd fd fd fd f9 fd fd fd f9 fd f9 fd f9 fd ùýýýýýùýýýùýùýùý f9 fd fd fd fd fd fd fd f9 fd fd fd f9 fd f9 fd ùýýýýýýýùýýýùýùý f9 fd fd fd f9 fd f9 fd f9 fd f9 fd fd fd fd fd ùýýýùýùýùýùýýýýý f9 fd f9 fd f9 fd f9 fd f9 fd f9 fd fd fd f9 fd ùýùýùýùýùýùýýýùý f9 fd fd fd f9 fd fd fd f9 fd f9 fd f9 fd f9 fd ùýýýùýýýùýùýùýùý fd fd fd fd fd fd fd fd f9 fd fd fd fd fd fd fd ýýýýýýýýùýýýýýýý f9 fd fd fd fd fd fd ff 42 01 0a 01 07 08 06 00 ùýýýýýýÿB....... 73 73 72 73 72 73 72 73 72 73 72 73 72 73 72 73 ssrsrsrsrsrsrsrs 72 73 72 73 72 73 72 73 72 73 72 73 72 73 72 73 rsrsrsrsrsrsrsrs 72 73 72 73 72 73 72 73 72 73 72 73 72 73 72 73 rsrsrsrsrsrsrsrs 72 73 72 73 72 73 72 73 72 73 72 73 72 73 72 73 rsrsrsrsrsrsrsrs 72 73 72 72 73 72 73 73 72 72 72 5f 5f f9 fd fd rsrrsrssrrr__ùýý fd fd fd f9 fd fd fd fd fd f9 fd f9 fd fd fd f9 ýýýùýýýýýùýùýýýù fd fd fd fd fd f9 fd f9 fd f9 fd f9 fd fd fd f9 ýýýýýùýùýùýùýýýù fd f9 fd fd fd fd fd fd fd f9 fd f9 fd f9 fd f9 ýùýýýýýýýùýùýùýù fd f9 fd f9 fd f9 fd fd fd f9 fd f9 fd f9 fd f9 ýùýùýùýýýùýùýùýù fd f9 fd f9 fd fd fd f9 fd f9 fd f9 fd f9 fd f9 ýùýùýýýùýùýùýùýù fd f9 fd f9 fd f9 fd fd fd f9 fd f9 fd f9 fd fd ýùýùýùýýýùýùýùýý fd f9 fd f9 fd f9 fd f9 fd fd fd f9 fd f9 fd f9 ýùýùýùýùýýýùýùýù fd fd fd fd fd fd fd f9 fd f9 fd fd fd fd fd fd ýýýýýýýùýùýýýýýý fd fd fd f9 fd fd fd f9 fd f9 fd f9 fd fd fd f9 ýýýùýýýùýùýùýýýù fd f9 fd f9 fd f9 fd f9 fd fd fd f9 fd f9 fd fd ýùýùýùýùýýýùýùýý fd f9 fd f9 fd f9 fd f9 fd f9 fd f9 fd f9 fd ff ýùýùýùýùýùýùýùýÿ 42 01 0a 01 47 08 86 00 73 72 73 72 73 72 73 72 B...G.†.srsrsrsr 73 72 73 72 73 72 73 72 73 72 73 72 73 72 73 72 srsrsrsrsrsrsrsr 73 72 73 72 73 72 73 72 73 72 73 72 73 72 73 72 srsrsrsrsrsrsrsr 73 72 73 72 73 72 73 72 73 72 73 72 73 72 73 72 srsrsrsrsrsrsrsr 73 72 73 72 72 73 73 72 73 72 73 72 72 72 73 72 srsrrssrsrsrrrsr 73 72 73 5f f9 fd f9 fd fd fd f9 fd fd fd fd fd srs_ùýùýýýùýýýýý f9 fd f9 fd f9 fd fd fd fd fd fd fd fd fd f9 fd ùýùýùýýýýýýýýýùý f9 fd fd fd f9 fd fd fd fd fd f9 fd fd fd f9 fd ùýýýùýýýýýùýýýùý f9 fd fd fd fd fd f9 fd f9 fd fd fd f9 fd f9 fd ùýýýýýùýùýýýùýùý fd fd f9 fd fd fd f9 fd fd fd f9 fd f9 fd f9 fd ýýùýýýùýýýùýùýùý fd fd f9 fd f9 fd f9 fd fd fd f9 fd f9 fd fd fd ýýùýùýùýýýùýùýýý f9 fd f9 fd fd fd f9 fd fd fd f9 fd f9 fd fd fd ùýùýýýùýýýùýùýýý f9 fd f9 fd f9 fd f9 fd fd fd f9 fd fd fd f9 fd ùýùýùýùýýýùýýýùý fd fd f9 fd f9 fd f9 fd f9 fd fd fd f9 fd f9 fd ýýùýùýùýùýýýùýùý fd fd f9 fd fd fd fd fd fd fd f9 fd f9 fd fd fd ýýùýýýýýýýùýùýýý f9 fd f9 fd f9 fd f9 fd f9 fd f9 fd f9 fd fd fd ùýùýùýùýùýùýùýýý fd fd f9 fd fd fd fd ff 42 01 0a 01 07 08 06 00 ýýùýýýýÿB.......

I saved the whole measurement (about 100 packages) as a txt File and imported it with Microsoft Excel . With the function “HEXINDEC” I’ve converted all bytes to decimal values. I sorted 2 different data packages (obviously type “BA”) by hand in an extra worksheet an ploted a graph:

Looks exactly like the “clean” and glitchy data on the PC Software. It seems like the “glitchy” signal is what is shown on the scopes display referring to the samples volt values. Just the high values are totally noisy.

Because I wanted to see all the data sent in a graph to see if there is a certain pattern or something I wrote a simple VBA code that extracts all sample data, ignoring the headers and writes it to another excel sheet.
Here is plot from all packages (about 100):

If you watch different time sections you can see how the “glitchy” and the “clean” signals alternate. And there are also samples that are zero.

In your excellent bulletin board I also found a description of the packages headers: http://forum.velleman.eu/viewtopic.php?f=9&t=1186
From over 100 packages in the text file there are only 6 starting with “BA”. I haven’t studied the header architecture in detail but I assume that not only the Samples are damaged but also the headers. This would explain why using the PC Software for the PPS10 new data is shown only every few seconds.

By the way: This post looks like a similar problem: http://forum.velleman.eu/viewtopic.php?f=9&t=2508. Just not to such a large extend as I am experiencing here. Is there any known solution ?

So after all these testing which cost me nearly half a day I came to the conclusion that my PPS10 has to be defect. Because I have the same results with different Hardware and different operating systems I am quite sure that the problem is not due to my configuration.

The only thing I haven’t tried is to use a external power supply instead of batteries. Any Chance this could change something because of higher voltage on the com port ???

As you can imagine I am really disappointed. I need the scope for my diploma thesis but I also want to use it furthermore for hobby purposes after the thesis. Because I study aerospace engineering and not electrical engineering, I have only basic knowledge in electronics. I hadn’t had high expectations to the device (never had a oscilloscope before). Most of the time I just want to measure some square wave pulses that are used on radio controlled vehicles.
But the PC Connection was the main reason I’ve decided for the PPS10. So I don’t have to take a picture of the display each time I want to show someone what I’ve measured and I am able to analyse data with other programs on the PC.
Otherwise I would have chosen the HPS140.

I guess the best thing to do would be to do the few measurements I need most urgent for my thesis and send the device pack to the dealer I’ve bought it just a week ago.
But I wanted to check if I have overlooked something or if there is any simple fix to the problem. Perhaps it’s a firmware problem, so sending it back would change nothing if the dealer’s devices have all the same firmware. In that case sending it directly to you would be better (but slower I guess).

On the other hand it would be good to have a confirmation that the problem lies definitely at the device, so the dealer can’t tell there would be a problem with my PC or something.

So I wanted to check here first.

Thank you very much for your advice in advance,
Chris

Very strange problem which has not been encountered before.
We suggest to return the unit for inspection/repair.
Please return it to your ditributor or to:

Velleman Projects Tech. Dept.
Legen Heirweg 33
9890 Gavere
Belgium

Make sure to include a copy of this conversation, so that our engineers know where to look.

Thank you for your advivce.
I’ve sent the package back to the distributor on Saturday. Unfortunately I got the last PPS10 they had. So they can’t send me another one as replacement and I have to wait for the repair.
Luckily I found out, that I can also use or even borrow an oscilloscope from the university. So at least there will be no delay for my thesis. Nevertheless I would appreciate a quick repair. But I guess if your engineers as quick as the Team here in the bulletin board that will be the case :wink:

Cheers