Hi
I Purchased a PCSGU250 yesterday.
I then installed the software
I then installed the unit and installed the driver software.
I downloaded the latest driver software from your site.
When starting the software the software started by stating it will calibrate.
It started calibrating 10% and then 20%.
Then it stopped there and I waited a long time but nothing Happened.
If not yet done, please download and install the latest driver “Drivers for Windows Vista and Windows 7. (Can also be used for Windows XP)” from: velleman.eu/support/download … U250&type=
Download and install also the latest software Rev 1.12 from the same site.
There may be hardware problem in the PCSGU250.
Can you try in other PC?
If not, you can try following procedure to check if the oscilloscope is working OK:
If you are running Windows Vista or newer PC, locate the file PCSGU250.ini (if exist) in the folder:
C:\Users<username>\AppData\Local\Velleman\PCSGU250
In older PCs the file is located in the folder: C:\Documents and Settings<username>\Local Settings\Application Data\Velleman\PCSGU250
Delete the file PCSGU250.ini
Disconnect the USB cable.
Run the PcLab2000LT.
Let it go to demo mode.
Then close the software.
Now locate again the file PCSGU250.ini and open it in a text editor.
In the file locate the line: HWdialog=1
Change it to: HWdialog=0
Save the file
Connect the USB cable and run the PcLab2000LT
In the Options menu select Hardware setup and there select PCSGU250.
Now the oscilloscope starts without the calibration and you can test its operation.
Connect the function generator output to the oscilloscope input to test.