MyPCS500 trigerring do not work. No one channal (CH1, CH2 or EXT), selected as trigger source, do not triggering. Calibration is all right, with result :
CH1 offset at 15V/div : 120 OK
CH2 offset at 15V/div : 119 OK
CH1 offset at 5V/div : 1 OK
CH2 offset at 5V/div : 2 OK
CH1 offset at 15mV/div : 0 OK
CH2 offset at 15mV/div : 3 OK
CH1 offset at 5mV/div : -2 OK
CH2 offset at 5mV/div : 10 OK
CH1 offset at 20us/div : 0 OK
CH2 offset at 20us/div : 0 OK
CH1 offset at 10us/div : -2 OK
CH2 offset at 10us/div : -2 OK
CH1 offset at 5us/div : -3 OK
CH2 offset at 5us/div : -5 OK
CH1 Trigger low : 102 OK
CH1 Trigger middle : 128 OK
CH1 Trigger high : 154 OK
CH2 Trigger low : 102 OK
CH2 Trigger middle : 128 OK
CH2 Trigger high : 152 OK
EXT Trigger middle : 104 OK
I’m tested my PCS500 at two computers, results are identical - triggering not work, but all other function work properly.
Device age 11 months
Does the triggering occur if you move the trace’s y-position slider up and down? (Across the triggering level mark.)
Does the triggering occur if you first put the free running trace about to the middle of the screen. Then put triggering on and then adjust the triggering level adjustment up and down across the middle of the screen?
I found anomaly :
When CH1 switch to 0.15V/Div (or GND and from 15V/Div to 50mV/Div) and select as trigger source then triggering is worked properly, and in this case EXT.TRIG is worked properly to.
When CH1 lived at 0.15V/Div (or GND and from 15V/Div to 50mV/Div) then triggering at CH2, work on whole diapason of V/Did. and Time/Div.
When Ch1 switch out from 0.15V/Div (or GND and from 15V/Div to 50mV/Div) position then triggering at CH2 stop working. If Ch1 switch back to 0.15V/Div and trigger switch OFF and ON, then triggering on Ch2 is worked again.