K8031 Transient adjustment and calibration problem

Dear Velleman staff & community members,

Recently i bought a k8031 oscilloscope from a local distributor. I assembled the device and tried to calibrate it. The offset adjustment works, however when i try to do the transient calibration, the scope fails.

Description of the problem:
I select 1V/div in pcLab and connect my probe to J4. The square wave that should be displayed does not fit on the screen, so I can’t adjust CV1. Even when I set RV2 to 0 ohm, I cannot get the waves to fit in the screen to calculate the rms value.

Checks done:

  • Checked RV2 resistance. It varies from 0 to 500 ohms as expected.
  • Checked signal ATT0-3, they all change when switching from 3V/div down to 10mV/div
  • Measured all the resistors in the analog part of the pcb to check for the correct value
  • Checked all the diodes for correct operation
  • Measured the voltage between probe ground and power ground -> 2.5V
  • Checked if RY1 to 5 open and close properly

Steps made to further diagone/resolve problem:

  • Removed RY5 from pcb and reconnected the inner two connections with a 2k resistor, this caused the wave to be displayed at 3V/div and 1V/div, however the minimum RMS at 1V/div is still 3.68V and at 3V/div 11.04V
  • I found that the resistance between the outer holes where RY5 resided, is only 2.1k, is this correct?

At this moment I really don’t know where to look for the error anymore. So I hope one of you has experience with this problem and know a way to solve it.

Thank you in advance!

Most probably there is something wrong with the input section.
Please check (once again):
R10 = 910k
R8 = 91k
R7 = 10k

RV5 seems to work OK.
It is closed on 1V/div range to increase the gain.
Now you removed it and the gain is equal on 3V/div and on 1V/div ranges.
The gain is too high due to some problem in the input section around R10, R8 and R7.

[quote]- I found that the resistance between the outer holes where RY5 resided, is only 2.1k, is this correct?[/quote]This is OK. There is R17 = 2k2 between the outer pins 1 and 7 of RY5.

Thank you VEL255, that did the trick,

It was R10, I removed it from the pcb and it measured 90k, although the resistor color code says it is 910k.
Until now i have never had such an error. The resistor colors are: white, brown, black, orange, brown. That should be 910k. Attached is a picture of the ‘anomaly’

I replaced it with an 910k resistor and the oscilloscope is working perfectly now. Thank you so much!

Greetings Sven

OK. I’m glad you found the problem and got the K8031 working!