Dear Velleman staff & community members,
Recently i bought a k8031 oscilloscope from a local distributor. I assembled the device and tried to calibrate it. The offset adjustment works, however when i try to do the transient calibration, the scope fails.
Description of the problem:
I select 1V/div in pcLab and connect my probe to J4. The square wave that should be displayed does not fit on the screen, so I can’t adjust CV1. Even when I set RV2 to 0 ohm, I cannot get the waves to fit in the screen to calculate the rms value.
Checks done:
- Checked RV2 resistance. It varies from 0 to 500 ohms as expected.
- Checked signal ATT0-3, they all change when switching from 3V/div down to 10mV/div
- Measured all the resistors in the analog part of the pcb to check for the correct value
- Checked all the diodes for correct operation
- Measured the voltage between probe ground and power ground -> 2.5V
- Checked if RY1 to 5 open and close properly
Steps made to further diagone/resolve problem:
- Removed RY5 from pcb and reconnected the inner two connections with a 2k resistor, this caused the wave to be displayed at 3V/div and 1V/div, however the minimum RMS at 1V/div is still 3.68V and at 3V/div 11.04V
- I found that the resistance between the outer holes where RY5 resided, is only 2.1k, is this correct?
At this moment I really don’t know where to look for the error anymore. So I hope one of you has experience with this problem and know a way to solve it.
Thank you in advance!