Please confirm the test signal from the back so I know I’m reading/using the HPS140 correctly …and to determine whether dementia and alzheimer’s have kicked in. Last time I used an oscilloscope was 30+years ago so my brain’s SNR is maybe -3dB at best.
I adjusted the probe capacitor in 10x mode with little plastic screwdriver to get a nice wave with sharp 90-degree corners.
75% duty cycle square wave (0.3 msec high / 0.1 msec low)
0.4 msec cycle = 2.5kHz frequency = 1 cycle/0.0004 seconds
My Volts peak-to-peak varies considerably by moving the X-position around, from 3.405 Vpp to 7.836 Vpp when viewing a trace stored in memory. How is this possible?
Am I doing something wrong? Vmin = -2.52 V, Vdc = 0.068V (it should be higher given the 75% duty cycle), and 1.396Vac.
Is the Vac accurate for every waveform, i.e. true RMS?
Looking at the “freq mark” my little Velleman is displaying 862.0Hz. Which doesn’t make sense.
The Fourier dude says I should have the fundamental frequency of the pulse train @ 2.5 kHz plus harmonics 5 kHz, 7.5 kHz, 10 kHz, 12.5 kHz, 15 kHz, 17.5 kHz, 20 kHz. If the waveform were symmetric I wouldn’t hear any even numbered harmonics, right? All of these are way more biggerer than 862 Hz.
I need to be able to trust the readings or at least understand where it’ll fail me and not rely on it while I’m up in a ceiling on a ladder over a french fryer with hot grease cooking onion rings below. My brain’s SNR drops to -110dB in such situations. I need to focus on avoiding to succumbing to 3rd degree burns and not performing mental math.
My parents wouldn’t let me do “math” in school because it’s a 4-letter word. Any assistance, even if half cocked and half vast, may have a modicum of utility. I’ll even appreciate it, so thank you ahead of time --for your time, …just in case I don’t.
Signal analysis should be intuitively obvious to the most casual observer and comprehension readily verifiable.