Hello,
I recently just purchased a HPS140 pocket scope to fine tune my car amplifiers, this is my first time using an oscilloscope so I had a couple questions…
I am going to be testing two amplifiers for clipping.
Amplifier A will be seeing voltages from 10-15 volts.
Amplifier B will be seeing voltages from 40-60 volts.
Should I be using the x1 probe setting for Amplifier A and x10 probe setting for amplifier B?
Also from what I am reading, once I change the probe setting to x10 all I would then need to do is go to settings and select x10? Once that is done am I good to go or is there anything else that I would need to do before I can test for clipping on the car amplifiers?
x1 should be fine, however, it is always best to start at x10 when measuring unknown signals.
The scope is equipped with a ‘x10’ setting, so readouts stay correct.
Scope can handle 100Vp at the input.
Probe can handle 600V at any range.
100Vp equals about 70Vrms for a sine wave.
Using ‘x1’ if possible results in higher resolution at low voltages.