saved images of software when running in three states (apologies, I got my numbering mixed up on the images):
- 2 - display and DSO file for screen when first turned on
- 1 - display and DSO file for screen when either +ve and/or -ve probes are toucing something (e.g. my finger)
- 3 - display and DSO file for screen when connected to 9v DC source
If I’ve missed anything here or if there’s anything else I can show that will help then do let me know.
The calibration values are OK.
Also the screenshots seem to be OK.
[quote]- 3 - display and DSO file for screen when connected to 9v DC source[/quote]Please select 5V/Div or 2V/Div range. Select DC coupling to measure DC source.
Please do not use Autoset for DC measurement.
Autoset turns always AC coupling on.
[quote]- 1 - display and DSO file for screen when either +ve and/or -ve probes are toucing something (e.g. my finger)[/quote]Please select Time/Div range 10ms to see some 50Hz waveform on the screen.
You should see something like this when touching the red probe with your finger:
When touching red probe with finger and selecting options as requested (range:2V/Div, DC coupling to measure DC source, Autoset off, time/div range 10ms) results in screen shot 4 (uploaded to the same Google Share as above).
I noticed on your screen shot that you had 0.1V/Div selected - trying this results in screen shot 5.
Indeed, doesn’t match. You may have very “low noise” environment.
Do you have any known signal source to test e.g. a little transformer with low voltage AC output.
Also you can check the operation with known DC sources e.g. batteries.
checking with several different batteries does not show any difference to the display. (Shows same as when nothing is connected or finger touching probe).
I have a Velleman MK105 Signal Generator (an earlier project) which also, when connected makes no difference to the signal displayed.
[quote]Could you tell me if there’s any particular specification I should be looking for when looking for replacement leads?[/quote]There are no special requirements. I think any flexible cable will do the job.
I’ve got a similar Problem, but I haven’t put the IC’s in with the notch to the false direction…
It passed Calibration, afterwards I connected it with the EDU06 and got a noisy line on 0,1 V (autoset…)
On the Pictures you can see one Picture with a small sinuswave - on this Picture i connected only the red cable which was connected to the IN-Port on the board…
Your EDU09 seems to be working.
There may be some problem in the input section.
What is the result if you connect the test leads to a battery (e.g. 1.5V) ?
Use DC coupling in this case.
I connected it to an 1,5V AA-battery - as you said to me - but I got almost the same Pictures. It showed me again a voltage about 0,0V-0,1V although the battery was new