I have assembled the EDU09 oscilloscoop, but the calibration failed. In the calibration log I found:
Calibration results:
CH1 offset at 5V/div : 161 OK
CH1 offset at 1V/div : 201 FAIL
CH1 Y-position low 32
CH1 Y-position mid 67
CH1 Y-position high 67
What did I check:
Check T1, T2 and T3
voltage over the zener diode ZD1: about 5V
E (voltage C30)= -2.6V
F (voltage C31)= -5.2V
Corrected the orientation of the 8 pin IC : was placed the wrong way around because of the missing notch.
In other post I read about no connection between C30 and C31. But how to check?
The charge pump seems to work OK.
The problem is IC2 (the 8 pin IC).
Please check the voltage on pin 1 of IC2, should be about 1.65V
On pin 7 there should be about -1.65V.
On pins 2, 3, 5 and 6 should be 0.0V.
CH1 Y-position low 24
CH1 Y-position mid 24
CH1 Y-position high 24
I checked IC2 as that’s said above. And I measured pin 8 and 4. I found 10,30 V. But, I think the good must be around 5V. What about you ? Which voltage did you find regarding pin 8 and 4 ?
I checked IC2 as that’s said above. And I measured pin 8 and 4. I found 10,30 V. But, I think the good must be around 5V. What about you ? Which voltage did you find regarding pin 8 and 4 ?[/quote]
Here’s my result during the checking:
pin 1: around 3,3 V
pin 7: 1,74 V (between 5 and 7)
-2,55 V (between 6 and 7)
between 2 and 3 = 4,32 V
between 5 and 6 = 0,24 V
It seems the voltage on pin 1 of IC2 is too high.
Have you checked the resistor values: R18, R19, R23, R25, R27, R9 and R10 ?
Here are the voltages on the pins of IC2 to GND (e.g. SK2).
The PC program is not running when measured.
IC2 pin 1: 1.64V
IC2 pin 2: 0.00V
IC2 pin 3: 0.00V
IC2 pin 4: -5.1V
IC2 pin 5: 0.00V
IC2 pin 6: 0.00V
IC2 pin 7: -1.63V
IC2 pin 8: 5.05 V
There are some strange results.
You measured -5.12V on pin 7 and -0.05 on pin 2.
The resistor R23 is between the pins 7 and 2.
You measured 1,51V over R23. This is very strange result (should be about 5 V).
There ais also other strange result:
Pin 5 is directly connected to GND.
You measured -0,23V between the pin 5 and GND - this should not be possible!
Can you please measure once again all the voltages on all pins of IC2 to GND.
[quote]I’ve put the red probe on the pins and the black on the GND.[/quote]This is OK.
A good GND is the shield of the USB connector or SK2 terminal of the input cable.
Otherwise, I have a question regarding L1. The first time I’ve switched on the kit, it was burning a little. So, there are some burn. But, the kit work. (I think that problem overcame because i connected IC2 in the wrong side). So, its voltage is around 5V, right ? (What I found L1 = 5,11V Both sides of its pins and what i saw according to the pictoral diagram from page 17).
[quote]So, its voltage is around 5V, right ? (What I found L1 = 5,11V Both sides of its pins and what i saw according to the pictoral diagram from page 17).[/quote]Yes, the L1 seems to be OK.
[quote]I got the problem. that’s R3, No voltage. Even if according the ohm test, the component is okay.[/quote]Seems to be OK too. There should be no voltage over R3.