[quote=“wooders”]I have carried out a few tests using a Fluke 45 meter (calibrated 7 May 2009), which has a Min/Max function that is useful for measuring the triangular waveform.
The pulse waveform used was created in PCLab2000LT rev 1.04 and is specified as follows:
1.0(1000)
-1.0(3000)The tests were carried out with the following settings:
Fine Tune settings set to their centre position.
Peak Offset and Auto Offset set to Off.
Frequency set to 16.5 mHz
Load >= 180Kohm
The test results are as follows:
Triangle Square 25% Pulse
Offset Amplitude min max min max min max
5.00 10.00 0.253 10.108 0.099 9.998 0.317 10.172
2.52 5.00 0.064 5.059 0.025 5.028 0.179 5.009
0.00 5.00 -2.489 2.507 -2.528 2.476 -2.374 2.453
0.00 10.00 -4.811 5.051 -4.972 4.927 -4.668 4.889This illustrates that the min and max are distinctly different for the 3 waveforms.
I have not carried out any stability tests nor any frequency checks (I’m not sure that I have anything available to test 16.5 mHz) so I am assuming the best.[/quote]
The Fluke 45 is a fine instrument, however it is a multimeter not a wave analyzer, with a maximum sample rate of 20 S/s and a minimum rate of 2.5 S/s. I believe the sluggish sampling rate may be coloring your observations, even with a 16.5 mHz signal.
The square wave remains at it’s peaks for 30.3 seconds, allowing plenty of time for the meter to sample and record the minimum and maximum values. However in the instance of the triangle wave unless the meter’s sampling occurred exactly in synchronisation with the signal’s peaks it would not capture a true peak reading.
Sampling the signal over some long duration would tend to allow the relation between the sampling rate and waveform repetition rate to perhaps “match up” for one or more samples, however with a 60.6 s period this could take hours.
This does not perhaps explain the 25% duty cycle issues, however there may be less obvious similar conflicts at work. I have checked the sine, triangle, and square wave function of my PCGU1000 and found the the sine and triangle amplitudes are dead on (4.96V p-p set to 5.0V), however the square wave exhibits a bit of over and under shoot (even when properly terminated), in the order of 1.25%, fully accounting for the slightly higher p-p measurements of 5.04V or so.